An Image/Link below is provided (as is) to download presentation Download Policy: Content on the Website is provided to you AS IS for your information and personal use and may not be sold / licensed / shared on other websites without getting consent from its author. Content is provided to you AS IS for your information and personal use only. Download presentation by click this link. While downloading, if for some reason you are not able to download a presentation, the publisher may have deleted the file from their server. During download, if you can't get a presentation, the file might be deleted by the publisher.

E N D

Presentation Transcript


  1. POC Leadership Forum Update on the Regulations CLIA, CAP, JCAHO and COLA Ronald H. Laessig, Ph.D. University of Wisconsin Madison, Wisconsin www.cms.hhs.gov/clia

  2. Purpose of U.S. Lab Regulations … quality standards for all laboratory testing to ensure accuracy reliability timeliness of patient test results regardless of where the test is performed… www.cms.hhs.gov/clia

  3. Series of six slides from Glamour Magazine The latest information on test quality Believe it or not #1 www.cms.hhs.gov/clia

  4. Series of six slides from Glamour Magazine The latest information on test quality Believe it or not #2 www.cms.hhs.gov/clia

  5. Series of six slides from Glamour Magazine The latest information on test quality Believe it or not #3 www.cms.hhs.gov/clia

  6. Series of six slides from Glamour Magazine The latest information on test quality Believe it or not #4 www.cms.hhs.gov/clia

  7. Series of six slides from Glamour Magazine The latest information on test quality Believe it or not #5 www.cms.hhs.gov/clia

  8. Series of six slides from Glamour Magazine The latest information on test quality Believe it or not #6 www.cms.hhs.gov/clia

  9. Quality POCT: What drives the system? According to Glamour Magazine: ….(we always go right to the source) If Regulations are not followed…... ...Human error follows J. Yost, the “CLIA Lady”

  10. Manufacturers’ Point of View Evolution in Assured Quality • Reduction in size • Operator ID / Patient ID • Reagent viability • Lock-out quality control • Connectivity • Data management Roche Diagnostics

  11. Regulator’s Point of View U.S. Regulations = CLIA C = Clinical L = Laboratory I = Improvement A = Amendments Regulates ~180,000 U.S. test sites http://www.cms.hhs.gov/clia/

  12. Regulator’s Point of View CLIA’88 (February 28, 1992) CLIA’03 (January 24, 2003)

  13. Linking Quality and Regulations CLIA’s working premise: CLIA’s mandated quality practices will improve quality of test results

  14. CLIA Regulations and the Director The laboratory director is responsible for the overall operation and administration of the laboratory … [the] testing systems … used [must] provide quality laboratory services for all aspects of test performance … including the pre-analytic, analytic, and post-analytic phases of testing

  15. Quality POCT Challenge – March 2004 Dilemma for today: • With the manufacturer building-in/assuring quality, • Where is the professional input from the laboratory director? • Does the director just buy the technology and use it?

  16. What do we know today about CLIA and Quality Requirements? • CLIA’03 takes a Quality Systems approach like ISO documents • The quality requirements follow the route of the specimen • pre-analytical • analytical • post-analytical issues

  17. CLIA and Quality Requirements? • The old CLIA covered the analytical requirements in one section and quality assurance in another. • The new CLIA groups the quality requirements under a new term “Quality Assessment” • Quality Assessment = Quality Control + Quality Assurance • CLIA wants to consider the testing process start to finish-hence “pre, post and analytical assessment”

  18. CLIA ‘03 and QC? Control procedures must (§493.1256): • Monitor the accuracy and precision of the complete analytical process • Detect immediate errors due to test system failure, adverse environmental conditions, and operator performance • Monitor over time precision and accuracy that may be influenced by changes in test system, environment and operator performance

  19. CLIA’03 and QC? Control procedures must (§493.1256): • Monitor complete analytical process • Detect immediate errors • Monitor over time precision and accuracy Conceptually, meeting the requirements sounds easy, But ---

  20. CLIA’03 and QC? Control procedures must (§493.1256): • Monitor complete analytical process • Detect immediate errors • Monitor over time precision and accuracy Conceptually, meeting the requirements sounds easy, But ---

  21. CLIA ’03 and Q C The Devil is in the Details

  22. CLIA’03 – The details! • What about • Electronic QC? • Process controls? • Internal controls? • Do these • evaluate complete analytical process • detect immediate errors • monitor accuracy and precision over time • none of the above

  23. Interpretive Guidelines for CLIA – 300+ pages NEW and More Exciting! Details - 12 January 2004 http://www.cms.hhs.gov/clia/appendc.asp

  24. New Concept for QC? EQC Equivalent Quality Control

  25. EQC = Electronic Quality Control New Concept for QC?

  26. What is Equivalent QC (EQC)? • For instruments with procedural/internal controls, test sites can choose to: • Analyze 2 external, liquid controls per day OR • Qualify the procedural/internal controls as one of three EQC options

  27. EQC Option #1 • Test Systems with internal/procedural control(s) that monitor the entire analytic process • Test site may use EQC, if test system’s stability is demonstrated

  28. EQC #1 - Evaluation of Stability • Along with internal/process controls, run 2 external QC daily for 10 consecutive days (why 10?) • If internal/process and external control results are acceptable (what is acceptable?) • reduce external QC from daily to once / month unless manufacturer requires more • Lab must perform/monitor the internal control(s) in accordance with manufacturer instructions

  29. Option #1 Instrument with manufacturer supplied on board controls, run automatically.

  30. EQC Option #2 • Test Systems with internal/procedural control(s) that monitor a portion of the analytic process • Test site may use EQC, if test system’s stability is demonstrated

  31. Option #2 - Evaluation of Stability • Along with internal process controls, run 2 external QC daily for 30 consecutive days (why 30?) • If internal and external control results are acceptable (what is acceptable?) • reduce external QC from daily to once / week unless manufacturer requires more • Lab must perform/monitor the internal control(s) in accordance with manufacturer instructions

  32. Option #2 Hand held instrument currently advertized as having on board, electronic controls.

  33. EQC #3 • Test Systems without Internal/Procedural Control(s) • Advancements in technology have produced systems • Capable of maintaining stable performance specifications over time • Minimally influenced by adverse environmental conditions and operator variance. • Test site may use EQC, if test system’s stability is demonstrated

  34. Option #3 - Evaluation of Stability • Along with internal process controls, run 2 external QC daily for 60 consecutive days (why 60?) • If internal and external control results are acceptable (what is acceptable?) • reduce testing external QC from daily to once / week unless manufacturer requires more • Lab must perform/monitor the internal control(s) in accordance with manufacturer instructions

  35. Option #3? ? What belongs in this picture?

  36. Option #3? Every Instrument! Vitros Hitachi aca, SMAC, etc. Prove stability for 60 days, run external QC once per week

  37. CLIA – “Equivalent” QC (EQC) “The director must consider the laboratory’s clinical and legal responsibility for providing accurate and reliable patient test results versus the cost implications of reducing the QC testing frequency.” EQC is a choice!

  38. Uncle Sam Has Uncle Sam gone Crazy?

  39. Uncle Sam Has Uncle Sam gone Crazy? • Maybe or Maybe Not!

  40. March 2004 Does EQC and the U.S. regulations open a new era for new technologies or does EQC just cause confusion?

  41. Quality POCT: What Drives the System? • To me, CLIA QC regulations say – • Continue the “old” way – 2 external QC/day • Qualify the test system under EQC options and reduce frequency of external QC

  42. Quality POCT: What Drives the System? • To me, CLIA QC regulations say – • Continue the “old” way – 2 external QC/day • Qualify the test system under EQC options and reduce frequency of external QC • EQC Option #3 ( for a test system with no internal/process controls) • Do you want your testing done on a system that qualifies under this option?

  43. The Devil is in the details • EQC Qualified System – External QC analyzed once per month or once per week. • EQC-1 EQC-2&3

  44. “The Devil is in the Details” §493.1282 Standard: Corrective actions …[if] external QC [results]…fail to meet the laboratory's established criteria for acceptability, all patient test results obtained…since the last acceptable test run must be evaluated to determine if patient test results have been adversely affected…

  45. Quality POCT: Drivers? • Manufacturers – Building the perfect instrument to meet the needs of the testing environment

  46. CLIA’s new QC/EQC Requirements • Have a two-year phase in • Inspection • Citations • NO punitive activities

  47. Manufacturers have “EQ”Options #1, 2, & 3 To be competitive in the marketplace, options #2 and #3 DON’T count

  48. The bottom line…..the “Regs” • We are in a two year phase in • We are all trying to figure out what they mean • At the end of the day, the laboratory director must make the call -- QC or EQC (1,2,or 3) • We are after all in the patient care business • Legally, Morally, ethically, technically, the patients (must) always come first.

  49. CLIA Regulations and the Director The laboratory director is responsible for the overall operation and administration of the laboratory … [the] testing systems … used [must] provide quality laboratory services for all aspects of test performance … including the pre-analytic, analytic, and post-analytic phases of testing